MT Capacitive Level Meter - Zero Point Calibration Operation Video Guide
Watch the video tutorial
Scan the QR code on the left with WeChat or your browser to watch the related video.
Scan the QR code on the left
Deleetech MT capacitive level meter zero point calibration operation guide: Place the induction pen at the 9 o'clock position and remove it when P0 is displayed; place the induction pen at the 9 o'clock position again and remove it when P1 is displayed. Place the induction pen at the 3 o'clock position to enter the P1 menu, place it at the 3 o'clock position again, and wait for the numbers to flash, indicating the zero point calibration is successful. Then place the induction pen at the 9 o'clock position, remove it when P1 is displayed, wait 10 seconds for automatic exit, and the zero point calibration is complete.