Deleetech DCT8051 Capacitive Level Transmitter Increment Modification Operation Guide: 
Place the sensing pen continuously at the 9 o'clock position, and when the screen displays P0, remove the sensing pen. Place the sensing pen again at the 9 o'clock position to switch to P1, then continue operating to enter P3. P3 is the increment modification menu. 
Place the sensing pen continuously at the 3 o'clock position to enter the P3 parameter editing interface. The current increment value is 89.0, with the target modification being 100. The flashing decimal 0 at the end does not need to be changed. 
Lightly touch the 9 o'clock direction once to complete the shift; the digit “9” will flash. Lightly touch the 3 o'clock direction once to change the flashing digit “9” to “0”. 
Lightly touch the 9 o'clock direction again to shift to select the tens digit “8”, then lightly touch the 3 o'clock direction to change “8” to “0”. 
Continue lightly touching the 9 o'clock direction to shift to the hundreds digit, and lightly touch the 3 o'clock direction to change the hundreds digit “0” to “1”.
After completing the parameter modification, place the sensing pen continuously at the 9 o'clock position to exit the P3 menu. Wait for the 10-second countdown to end, and the increment parameter modification is complete.