MTM-A Capacitive Level Meter - Zero Calibration Guide After Interface Upgrade
Watch the video tutorial
Scan the QR code on the left with WeChat or your browser to watch the related video.
Scan the QR code on the left
MTM-A Quality Type Level Meter (after interface upgrade) Zero Calibration Operation Guide: Place the sensing pen at the 9 o'clock position and keep it still, then wait until the screen displays P0 before removing it. Lightly tap the 9 o'clock position once to switch the interface from P0 to P1. Place the sensing pen at the 3 o'clock position and keep it still to enter the P1 zero calibration interface, waiting for the values to stabilise. Place the sensing pen at the 3 o'clock position again and keep it still to perform the zero calibration; the digits flashing indicates successful calibration. Place the sensing pen at the 9 o'clock position and keep it still to exit the P1 interface. Wait for the 10-second countdown to finish; zero calibration is complete.