8051 Instrument Full-Scale Calibration Operation Video Guide

Jiangsu Deleetech 8051 Instrument Full-Scale Calibration Operation Guide: Place the induction pen at the nine o'clock position on the instrument until the interface displays P0, then remove the induction pen. Place the induction pen at the nine o'clock position again and adjust parameter P0 to P2, then remove the induction pen once completed. Place the induction pen at the three o'clock position on the instrument to enter the P2 full-scale calibration interface and then remove the pen, waiting for the interface values to stabilise. Once the values are stable, place the induction pen at the three o'clock position to carry out the full-scale calibration operation. When the interface values flash, it indicates calibration is complete, then remove the penetration pen. Finally, place the induction pen at the nine o'clock position to exit the P2 calibration interface and remove it, waiting for the 10-second countdown to finish. The 8051 instrument full-scale calibration operation is now fully complete.